Agilent GCMS Model 7890A/5975C vs 7890A/220 Ion Trap

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
Hi:

My company is looking to buy GCMS Model 7890A/5975C or 7890A/220 Ion Trap. If anyone has used this Agilent 7890/220 Ion trap model and please share your experience. Agilent claims it is almost equivalent to GC/MS/MS. Is it worth it?

My company will be analyzing cosmetic products such as lipstick, mascara, shampoo, conditioner, detergents, fragrances etc. Focusing on quantitation of Diethylene glycol, 1,4 dioxane, allergens, pescticides in health supplements etc. I found this forum is very helpful for a starter. Thanks

Is Perkin Elmer is better? Or Shimadzu?

I searched for this topic in the forum but I could not find. Please apologize for starting the new forum.
I would worry about using an ion trap for quantitation, a SQ or TQ would likely be better.

As for which manufacturer is better, this is a very subjective thing.

Best bet to give a sample to each one for a demo and see what they come back with. You'll get a feel for which company you like best as well.

I'm obviously biased towards Bruker (see signature), so don't overlook them in favour of the 'big boys'.

Where are you based?
Bruker (and Varian) Engineer happy to give advice on GC, GCMS and LCMS products.
Dear Federism

Do you have experince of Bruker GCMS. If yes, please share with us in detail as compared to Agilent.
He works for them :wink: If he is ex Varian then he may know the 220 well too.

His advice agrees with mine. Contact all the reps, get samples run and ask for local references. There is no substitute for good local support.
Having owned and operated 3 Varian 2000 series ion traps, I recommend a SQ for your application. The Ion Trap can do MS/MS which in certain applications is highly advantageous but doubt you need it. Developing methods on an ion trap is much more difficult as the method needs to be designed around the matrix. We used our ion traps to test about 60,000 oil samples for PCBs and the method worked very well but have since moved to TQ instruments.

I have a Bruker Scion and find it excellent.

Mark
Regarding your analytes and matrixes, you should have gc-ms-ms not just a single mass.
6 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry