PAHs by LC-MS/MS

Posted: Thu Jan 02, 2014 8:35 pm
by hplc.chem
Hi All,
Is it possible/advantageous to use LC-MS/MS for the analysis of PAHs?
Is there any standard method using LC-MS/MS for PAHs?

Thanks in advance!

Re: PAHs by LC-MS/MS

Posted: Fri Jan 03, 2014 3:36 am
by JMB
Possible ?---yes (MAYBE). Advantageous ?---NO.

PAH's analyzed by LC/MS (ESI or APCI) give very stable [M+H]+ ions---difficult to fragment, especially to differentiate the isomers.

You might research APPI (atmospheric pressure photo-ionization) to see if that is more promising than ESI/APCI.

However, why re-invent the wheel ?---there are well-established EI/GC/MS methods, which have the advantage of very high plate counts of capillary columns to resolve the isomeric PAH's. There is also the matter of identification from an EI spectrum; there are commercial databases (NIST etc) for library search that are sold with GC/MS systems.

Re: PAHs by LC-MS/MS

Posted: Sun Jan 05, 2014 3:36 am
by hplc.chem
Thanks for your reply.
Actually, our GC-MS system is not behaving well with high molecular weight PAHs!! :wink:

Re: PAHs by LC-MS/MS

Posted: Tue Jan 07, 2014 4:34 pm
by James_Ball
Thanks for your reply.
Actually, our GC-MS system is not behaving well with high molecular weight PAHs!! :wink:
If it is loss of linearity have you tried the larger draw out plate for the ion source?(Assuming it is an Agilent)

Re: PAHs by LC-MS/MS

Posted: Wed Jan 08, 2014 4:42 am
by hplc.chem
Thanks.
It is an Ion Trap GC-MS. I get dirty chromatogram at temperatures around 300 oC and above.