Help with Gradients

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi all - Can anyone recommend some good online references or tutorials for learning about gradient use in reverse phase HPLC? Specifically, in what way do changes to the gradient effect the resulting chromatography? I am a biologist who has been thrust into the world of analytical chemistry, and I am having difficulty grasping the nuances of gradients, and I am forever guessing wrong about what adjustments to make. For instance, if I want to move my peak to a later elution time, I'll make whatever change I think is necessary, and it either makes no difference, or moves it the opposite direction! Or if I need to sharpen up a peak, I don't know whether to flatten out the gradient or make it steeper, or where to do this in the timing. The chemists here are able to tell me where, when and how to accomplish things, but it's difficult in the hurry-up world of work-flow to get a detailed explanation of WHY it works. Can anybody help?

Thanks.
I am assuming you have some knowledge of isocratic type elutions, where you have a constant mobile phase throughout your anaylsis.

In isocratic methods, a higher percent organic content (referred to as %B from here on out) will lead to sooner elution time. Likewise, if you decrease the %B, you will have a longer elution time.

Gradient elution is pretty similar in some ways. Instead of changing %B though, you change your %B per minute. For instance, if you increase your %B/min, your peaks will elute sooner. If you decrease your %B per minute, your peaks will elute later. Higher %B/min gradients lead to sharper peaks, but will also push any peaks you have closer together.

One important thing to note is that every HPLC system has a "void volume". In an isocratic method, this isn't as important as in gradient methods. You can typically see the period of time it takes for the void volume to be run through by a dip in the baseline of your chromatogram. This is the time that the injection starts to get to your detector. Data handling systems typically record your chromatogram starting at the time your sample is injected in the system.

Let's say that the time for the void volume to pass is 2 minutes. That means the initial conditions of your gradient are hitting the detector just then. Even if you have your instrument starting at 0%B and going at 10%B/min, that portion of the gradient isn't hitting the detector until 2 minutes because of the 2min void volume. With this knowledge, you can estimate approximately what %B range your peaks are eluting. An example of this is, assuming you have a 2 minute void volume and a gradient that is 10%B per minute, if you have a peak eluting from 4-5 minutes, that peak is actually being eluted at 20-30%B because the 2 minute void volume does not have the gradient running through it.

IF you determine that your peaks of interest are coming out from, for instance, 30%B to 50%B you can clip your gradient to run from perhaps 20% to 60% so your gradient is not as long.

In simpler terms for gradients:
Adjusting %B/min higher will lead to quicker eluting, sharper peaks but could lead to bad resolution between peaks.
Adjusting %B/min lower will lead to slower eluting, less sharp peaks but could lead to better resolution between peaks.
Determine your void volume visually if possible. This is the point where your gradient is actually starting (if your instrument method does not have an isocratic portion at the beginning). This can help you to know at about what %B your compounds are eluting.

This is a pretty basic view of gradient type elutions. I am not completely sure of any on line resources and will let others add to this.
Even in our fast-paced, internet-driven world it's sometimes nice to read a good old-fashioned book :D. See if you can get your hands on one of the following:
- The bible: "Practical HPLC method development" by Snyder, Kirkland and Glajch
- An update, focused on gradients: "High performance gradient elution" by Snyder and Dolan

There were some articles by Dolan on LCGC online last year, maybe you want to have a look on these, too:
http://www.chromatographyonline.com/lcg ... oryId=9678
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry