film thickness calculation in capillary column
Posted: Wed Dec 25, 2013 3:48 pm
by jdezeeuw
Hi
I am looking for reference for actual calculation of the average filmthickness in WCOT capillary columns used for GC. I know it can be done via the chromatography using the specific retention volume, the column ID, and phase density, but I need a reference for the formula.
Cannot find anything related on the web. Anybody out there that has this? Thanks!
jaap de Zeeuw
Re: film thickness calculation in capillary column
Posted: Sat Dec 28, 2013 3:49 am
by thohry
Hi
jaap de Zeeuw
You can do that with isothermal condition using the equation:
k = K/beta
where
k is the retention factor
K is distribution coefficient,
beta is the phase ratio which is = (gas phase volume)/(stationary phase volume).
Posted: Sat Dec 28, 2013 10:21 am
by jdezeeuw
Thank you, this is possible. The challenge is that the K is depends on the liquid phase characteristics which vary a lot. Also there are very few reliable literature values for K. Problem is that the GC phases are not 100% defined. OV-101 has different K then OV-1 although both are 100% polysiloxanes. If you look at bonded phases the story becomes more complex. If we include polar phases it becomes even worse.
I am looking for the formula that has the specific retention volumes.
regards
jaap
Posted: Sun Dec 29, 2013 9:45 am
by Peter Apps
Hi Jaap
I suspect that this is buried deep in the packed column literature - the phase volume in the column has the same effect on retention no matter the column geometry - probably in one of the early books.
Happy new year
Posted: Sun Dec 29, 2013 1:28 pm
by jdezeeuw
yes, probably: I did a little math and the equation I came up with is this:
df = 0.5 x (r x k) / Vg x SD
where:
df = film thickness in um;
r = column radius in um;
k = retention factor reference component at test temp.;
Vg = specific retention volume for reference component at test temp. in cm3/g on the phase used
SD = specific density of phase in g/cm3;
so it's pretty easy. Now we only need a list of Vg values. I do have a few numbers we used and it fits pretty good
jaap