Good day..
I am newbies in GC world and have few doubts. Really hope for guidance from you all. Thanks.
Actually I've by GC column DB-5 (30m x 0.32mm x 0.5um) and I have try to install the column according to the procedure as below :
1.Column installation (Done-OK)
2. Leak and installation check
3. Column conditioning
4. Setting linear velocity / flow rate
5. Bleed profile
6. Test mix
However I am stuck at 2nd step.
I've try on direct inject of non retained compound (source:lighter) and there is no peak obtained
The test condition used as below:
GC instrument: PE 1100 autosystem
Carrier gas: Nitrogen
Liner: Wide bore liner
Mode: Split
Carrier gas pressure: Ive try with: 1) 2.6psi // split vent: 85.6ml/min, 2) 6.2 psi// split vent: 180ml/min, 3) 25.0 psi// split vent: 444ml/min
Oven temperature: 150 oC (isothermal)
Injector: 250 oC
Detector: FID (300 oC)
Running time: 30 mins
Carrir gas set at tank: 90 psi
Compared to condition as column performance cert.
Inlet: 250 oC split at 75 ml/min
Detector: FID (320oC)
Carrier gas: Hydrogen
Flow: 38.8 cm/sec (1.9 ml/min)
Hold up time compound: methane (1.288min)
temperature program: Isothermal at 150 oC
My questions are:
a) the split mention as per cert was 75ml/min , How to check the figure? Does the figure come from the detector or split vent using flow meter?
b) Does my split ratio was to small and all sample was vent out?
c) According to the Van deemter curve, N2 diffusion was lower than H2 and He. Do I need to double the figure of split or any parameter as mention on cert cause using N2 as carrier gas?
d) Could anyone suggest on the method for non retained compound using N2 as carrier gas?
Thank you.
Sofia Atiqah