does anyone use GC/FID shimadzu 14 A
Posted: Sun Aug 14, 2005 6:19 am
by nina
Hi
do you work at high senitivity?(analysis at the ppb range)
Posted: Tue Aug 30, 2005 12:25 pm
by dlorenzo
Hi, I work with 2 GC-14 B, sometimes at higher sensitivity. Please tell me what do you want to know
Regards
DL
GC
Posted: Tue Aug 30, 2005 1:02 pm
by nina
I work with GC shimadzu 14A (FID) equipped with split/splitless injector.
column is PEG.
I inject manually with 1ul syringe.((FAME) methyloleate), split ratio 1:50.
bu the results do not repeatible, and can not detect low concentration (below 1 ppm).
injector lines and detector are clean.
my column is new and conditioned.
carrier gas is helium.
please help me
have you the same problem?
thank you.
ppm FAME not seen
Posted: Tue Aug 30, 2005 3:06 pm
by chromatographer1
Your injector or detector temperature maybe too low.
Your column temperature may not focus your injection plug properly.
Your injection technique may not be suitable.
Your detector may have the wrong flows. Check with a known std for sensitivity of the detector.
Your actual split ratio may not be what you think it is.
good luck.
Re: ppm FAME not seen
Posted: Wed Aug 31, 2005 11:13 am
by nina
Your injector or detector temperature maybe too low.
Your column temperature may not focus your injection plug properly.
Your injection technique may not be suitable.
Your detector may have the wrong flows. Check with a known std for sensitivity of the detector.
Your actual split ratio may not be what you think it is.
good luck.
hello
injector temp: 230
detect temp:240
column temp:150C-5C/min-220
I check the flows, these are correct.
if sensitivity to be low,do I do?
thanks
Posted: Wed Aug 31, 2005 1:47 pm
by dlorenzo
Hi again, you must detect your sample at 1 ppm. I suppose you're using the 10^3 range, and working with a FID. Are the H2 and air flows in the optimum range?
Anyway, your can work with a low split ratio, i.e. 1:10.
BTW what is your sample's solvent?
good luck
MB
Posted: Wed Aug 31, 2005 2:46 pm
by nina
Hi again, you must detect your sample at 1 ppm. I suppose you're using the 10^3 range, and working with a FID. Are the H2 and air flows in the optimum range?
Anyway, your can work with a low split ratio, i.e. 1:10.
BTW what is your sample's solvent?
good luck
MB
thanks for reply
I work at 10^1 range, and FID.the concentration is 25 ppm but the peak is very small.
sample solvent is methanol.
I use standard addition for quantitation of FAME(oleic acid), is it correct?
and my resalts aren't repeatable.
best regards
low response
Posted: Wed Aug 31, 2005 3:47 pm
by chromatographer1
Did you check your detector sensitivity with a hydrocarbon std?
You must determine if the GC detector is working properly.
If it is then you must suspect your std.
If it is not, then you must repair the problem.
Good luck.
Posted: Tue Sep 20, 2005 10:30 pm
by Hafiz Allah Mehr
Hi, If your sample is oil and you do esterification to metylesters, you may have problem in your esterification technique, may not have two layers well seperated (add saturated NaCl solutin to make good separation) or incomplete esterification (use less amount of sample).