Evaporative LIght Scattering
Posted: Wed Dec 18, 2013 6:49 pm
by bdhedhy19
Hi all,
I have to develop the method using Evaporation Light Scattering (PL-ELS 2100 detector). I want to use DMF instead of THF, wonder if it is possible? if i increase the evaporation temperature or if you have any suggestion, please let me know.
thanks
Re: Evaporative LIght Scattering
Posted: Thu Dec 19, 2013 6:44 pm
by Consumer Products Guy
Hi all,
Yes. You likely will need to raise the nebulizer temperature.
Posted: Thu Dec 19, 2013 10:21 pm
by Vlad Orlovsky
Why do you want to use DMF? Any particular reason, like selectivity, solubility or else? I can't imagine a reason why you want to do this.
Posted: Sun Dec 22, 2013 6:23 am
by fchq2001
Hi all,
thanks
As I know the evaporation temperature is depends on the sample
Posted: Wed Jan 08, 2014 5:24 pm
by bdhedhy19
thanks for the tips with ELSD, really appreciate it
I used DMF for PMMA and polyurethane samples using Wyatt Instrument (Refractive Index) for another company, now i am working for different company and they have ELSD GPC Instrument.
I dissolved the samples and PMMA in DMF and it shows peaks, Now i am trying to adjust the Nebulizer temperature and evaporative temperature to get resolved and smooth peak.
Agilent recommended to use Ammonium acetate or Ammonium formate (0.1%) in DMFmobile phase. I will try to increase the Nebulizer temperature and will see.
Yes, Vlad Orlovsky I want to use DMF, because of the good solubility of most of the samples. Which solvent would you recommend for ELSD GPC.
Please let me know
Posted: Wed Jan 08, 2014 5:30 pm
by Vlad Orlovsky
from my experience a lot of compounds dissolve in EtOH/THF mixture. My guess is that with DMF you will need to set up temperature of ELSD around 100*C. Make sure that your compounds are not evaporating at this temperature becuase you will see poor linearity at lower concentrations.
Posted: Thu Jan 16, 2014 5:29 am
by bdhedhy19
Thanks Vlad for your suggestion:)
what ratio would you recommend for this THF:EtOH mixture (90% THF:10% EtOH) or what? Please let me know
when you have time.
I really appreciate it.
thanks so much