Hi every one,I am regular user of HPLC,Please suggest 10 or higher columns switching valve compatible with chemstation software

Re: multi column switching valve for agilent 1200 HPLC

Posted: Thu Oct 06, 2016 8:08 pm
by Multidimensional
You may want to consider purchasing an LC SPIDERLING™ HPLC COLUMN SELECTION SYSTEM for your Agilent brand HPLC system. They are available in room temp as well as fully heated / sub-amb cooled versions for up to 9-columns plus one By-Pass line and work with all versions of ChemStation for LC or OpenLAB (and other systems too).

http://www.hplctools.com/lcspiderling.htm

Re: multi column switching valve for agilent 1200 HPLC

Posted: Fri Oct 07, 2016 2:48 pm
by Consumer Products Guy
Wow - that's a lot. We just used the Agilent 2-column switching valve, that was plenty. 10 columns gets complicated with plumbing, etc.

I'd say stay simple and use fingertight fittings.

Re: multi column switching valve for agilent 1200 HPLC

Posted: Tue Oct 11, 2016 7:45 pm
by James_Ball
I remember there being a 6 column external valve made by Agilent but for more than that you had to go third party.

I once made a setup to run Glyphosate, Carbamates, and Diquat all on one system without removing columns but it was complicated. I used the two way valve in the Agilent column oven with one side being the Diquat column then the other passing out to a Pickering Post Column unit where I had another two way valve that switched between the Glyphosate and Carbamate columns then the common returned to the "IN" port of the two way in the column oven.

If you get creative you can make almost anything work :)