by
lmh » Mon Dec 16, 2013 4:53 pm
A lot depends on what is behind that requirement for 0.1unit mass resolution.
If what you mean is "I'd like an instrument that reads mass accurately to 0.1 units" then any modern quadrupole-based instrument will do the job. In my experience pure ion-traps are potentially a little less accurate, but in the same ball-park. Orbis and ToFs will, of course, do muuuuuuch better.
If you want to resolve peaks that are only 0.1 units apart, then a quadrupole won't do it, but an orbitrap is a bit of an overkill. Your application is uncomfortably between the two. There are traps with this sort of resolution (e.g. Bruker's trap, which claims a resolution of about 30,000 I think - completely stunning!), but this is resolution without accompanying mass-accuracy, and is really intended for people who want to resolve isotope peaks on proteins/peptides with very high charge-states. There are also ToFs that offer appropriate resolution, but they'll come with a price-ticket that's similar to an Orbitrap.
The next questions are (1) How sensitive do you want to go? and (2) Will you ever need full-scan qualitative work? If you need the ultimate in sensitivity, a triple quad has a lot to offer, but my limited experience is that super-sensitive triple quads are awful for general purpose full-scan work, where a trap will do much better. I haven't tried something like the Q-Exactive, so I have no idea where it would fit in.
Incidentally, the question on linear trap versus 3d trap is complicated. Thermo will tell you that linear is utterly the best (because it's what they do) and Bruker will tell you that 3d is best (because it's what they do). Both companies have tackled the problem of space-charging and capacity in the old-fashioned 3d traps, but in different ways, and the best thing is to get a good technical rep from each company to explain what they've done, and then test the rival instruments with some samples of your own. Good luck!