Bizarre change in HPLC results. Did I destroy my column?
Posted: Sat Dec 14, 2013 9:15 pm
Hey guys,
New to the forum, I appreciate any advice you might have for someone fairly new to HPLC. Unfortunately I don't have any data on this, my home computer, so I'll do my best to describe it. Otherwise I can attach chromatograms on Monday.
Here's the breakdown:
Analytes: Capreomycin sulfate, antibiotic #2 (AB2)
Column: brand new Zorbax Eclipse Plus C18, 3.5 um, 4.6x150 mm (NO GUARD COLUMN)
Mobile Phase A: 0.1% TFA in H2O (pH 1.95-2.05)
Mobile Phase B: 0.1% TFA in ACN
Problem: The first few runs were great. We had three peaks, one for capreomycin IIA/IIB, one small for capreomycin IA/IB, and one for AB2, all around expected elution times. Eventually I had problems with clogging somewhere in the sample delivery system (no peaks, the auotinjector leaked, and the injection port overflowed). I gave up for the day and washed the column with about 5 mL mobile phase B at the advice of my PI. The next day my PI helped me backwash the 6-way valve and we got peaks again. However, my capreomycin peaks were shifted about 15 minutes back with the AB2 peaks unaffected. I then did the following things:
1) tightened some loose junctions where mobile phase was leaking slightly
2) changed my mobile phase A since there was visible evaporation at the top of the bottle
3) changed my sample
From then on, I got the consistent result of four strange peaks, two of them broad, early in the run for capreomycin.
Do y'all have any idea what could be going on? Might the low pH of my mobile phase A have destroyed the column? What about backpressure from the earlier clog? Why would this affect ONLY the capreomycin and not the second antibiotic? I appreciate any insight or troubleshooting advice you could offer. We're probably ordering a new column and guard column on Monday just in case.
New to the forum, I appreciate any advice you might have for someone fairly new to HPLC. Unfortunately I don't have any data on this, my home computer, so I'll do my best to describe it. Otherwise I can attach chromatograms on Monday.
Here's the breakdown:
Analytes: Capreomycin sulfate, antibiotic #2 (AB2)
Column: brand new Zorbax Eclipse Plus C18, 3.5 um, 4.6x150 mm (NO GUARD COLUMN)
Mobile Phase A: 0.1% TFA in H2O (pH 1.95-2.05)
Mobile Phase B: 0.1% TFA in ACN
Problem: The first few runs were great. We had three peaks, one for capreomycin IIA/IIB, one small for capreomycin IA/IB, and one for AB2, all around expected elution times. Eventually I had problems with clogging somewhere in the sample delivery system (no peaks, the auotinjector leaked, and the injection port overflowed). I gave up for the day and washed the column with about 5 mL mobile phase B at the advice of my PI. The next day my PI helped me backwash the 6-way valve and we got peaks again. However, my capreomycin peaks were shifted about 15 minutes back with the AB2 peaks unaffected. I then did the following things:
1) tightened some loose junctions where mobile phase was leaking slightly
2) changed my mobile phase A since there was visible evaporation at the top of the bottle
3) changed my sample
From then on, I got the consistent result of four strange peaks, two of them broad, early in the run for capreomycin.
Do y'all have any idea what could be going on? Might the low pH of my mobile phase A have destroyed the column? What about backpressure from the earlier clog? Why would this affect ONLY the capreomycin and not the second antibiotic? I appreciate any insight or troubleshooting advice you could offer. We're probably ordering a new column and guard column on Monday just in case.