I’m having some trouble with a steadily dropping peak area over the course of a run. Assay details are as follows:
Equipment:
Waters e2695 separations module
Waters W2475 fluorescence detector
Mono Q 5/50 (IEC) or Superdex 200 10/300 (SEC)
PTFE sample tubes
Reagents:
Filtered Mobile Phase containing Salts, Tris Base and 10% Polysorbate-80 at PH7
Run Time:
40 or 30mins depending on the assay. Total run length, 30+hrs
Sample:
Protein (blood clotting factor)
I run two methods, a gradient IEC method and an isocratic SEC method. In both assays I see a drop in peak area throughout the run. To this end we have bracket injections throughout the run and drift injections at the end.
I have introduced a new reference standard* and have not managed to get beyond the first or second bracket injection before Peak Area becomes too low (system suit. agreement <97%). If I could figure out what aspect of the assays causes the drop in Peak Area I’m sure I could take steps to negate the effect. Any Ideas?
Note: My current best guess is maybe the PS-80 is oxidizing or there may be some other change in the Mobile Phase. I have tried using it fresh and leaving it 48hrs but not often enough to notice a pattern.
* The new Ref. standard has a lower protein concentration. This means that when reconstituted in 500µl, the sample is more dilute. We alter the on-column protein conc. by adjusting the injection volume accordingly.