sepearation of Lactate and acetate on PRP-X 100 column
Posted: Wed Dec 11, 2013 6:15 am
by hiren888
I would like to detect Lactate, Acetate and sulfate using PRP-X 100 column with indirect UV method. The problem is, I am not able to get separation of lactate and acetate. (mobilephase 4mM p-hydroxy benzoaic acid pH 10) I need some help in this regard.
thanks in advance
Re: sepearation of Lactate and acetate on PRP-X 100 column
Posted: Wed Dec 11, 2013 4:01 pm
by tom jupille
Try decreasing the pH.
Posted: Wed Dec 11, 2013 4:41 pm
by Markus Laeubli, Metrohm
There is a pretty old paper:
The determination of lactate, acetate, chloride and phosphate in an intravenous solution by non-suppressed ion chromatography; Haddad,P.R. Croft,M.Y.; Chromatographia,
1986 /
21 / 11, P. 648 -650.
Eluent: 0.3 mM phthalate (pH=6.0), 30% ACN
Acetate and lactate are well separated within 9 min. But no information about sulfate. Iassume that ic comes pretty late.
The paper shows as well that acetate/lactate separation strongly improves from co-elution at 0% to approx. 4 min. RT difference at 50 % ACN.
You might try ACN with your eluent as well, may be with pH change as well.
Posted: Sat Dec 14, 2013 5:49 am
by DJ
Use a lower concentration of phthalate(1 mM) and add some MeCN (5-15%).
Bear in mind that conditions that facilitate separation of these early eluting organic acids will likely lead to excessively long retention times for sulfate. You may consider a step gradient to 10 or so mM phthalate once acetate and lactate elute in order to get sulfate off in a reasonable time.
Posted: Tue Dec 17, 2013 5:46 am
by hiren888
Thank you very much for nice suggestion, will try and get back to you.
Thank again
Posted: Mon Jan 13, 2014 11:02 pm
by ph/Amr Tarek
Separation of lactate and acetate by HPLC :
- mobile phase : 1000 ml distilled water containing 1.5 ml phosphoric acid .
- column : 150 × 4.6 mm , C18
- detector : UV 210 nm
- injection volume : 20 micro litre
- column temp : 25
- flow rate : 1 ml/min
I am applying this simple method usually and it gives very good results
ph/Amr Tarek
Instrumental Unit Head at Otsuka pharmaceuticals
Posted: Mon Feb 17, 2014 6:49 am
by hiren888
Thank you very much for nice suggestion. will try one by one and get back to forum if get some success.
Hiren