This helps. OK, your original posting asked about the possibility of dissolving ammonium acetate in pure acetonitrile to prepare a mobile phase. That's not what this article is promoting, though, which is why we were all confused. The article describes a gradient of 20-55% ACN in 0.1% ammonium acetate with a reversed-phase (ODS) column or a 20-32% ACN gradient in 0.1% ammonium acetate with a cyanopropyl column ("reversed-phase lite"). I suspect you assumed that you'd have to start with one mobile phase consisting of purely aqueous 0.1% ammonium acetate, a second mobile phase consisting of pure ACN with 0.1% ammonium acetate, and that you'd generate the gradient by blending them online. Not so! If I were you, I'd do the following:
1) Prepare a stock solution (say, 500 ml) of 1% (w/v) of ammonium acetate in water. Do this by weighing 5 g. of the salt into a 500-ml volumetric flask. Add about 250-300 ml of water and swirl until the salt is dissolved. Add enough water to reach the calibration mark and invert 8-9x. Use an HPLC grade of the salt if available, and filter the stock solution (0.45 um nylon filter) if the water wasn't already filtered and if you aren't using HPLC-grade ammonium acetate.
2) To prepare 1 L. of a mobile phase containing 20% ACN with 0.1% ammonium acetate, put 100 ml. of that stock solution into a 1-L. volumetric flask. Add 700 ml. water. Now add enough ACN so the volume is about 1-2 ml below the calibration mark. Invert 8-9x. The solution will cool. Let it warm up to room temperature (hint: I like to sonicate the solutions in a sonicator bath filled with warm or hot water for 1-2 minutes). The solution will expand (which is why you don't add ACN to the calibration mark at the beginning). Now add enough ACN to reach the calibration mark. Invert 8-9x.
3) Use the same method to prepare the other mobile phases with different % ACN, keeping the volume of added stock solution the same but varying the volume of additional water and ACN.
Hope this helps.
Dear Andy,
Thanks a lot for your proper and accurate description. It is really helpful. As I am new in HPLC (specially gradient one), I thought that I have to prepare the pure ACN with 0.1%Ammonium acetate as a source of mobile phase and HPLC machine can make the gradient 20-55% solvent for gradient elution. (Since I have done an isocratic run with pure ACN and 0.1%TFA as a source of the mobile phase (A) and water with 0.1 TFA (B) then machine made the 80% mobile phase (A) by itself during the run).
If I have to prepare the gradient mobile phase before run, I don't know which percentage of mobile phase in the range of 20-55% I should prepare, Just 20% and 55%? If so, Shall I put them as A and B? So Water with 0.1% ammonium acetate would be in C?