Well, it appears I was overly optimistic in thinking I could detect CO with a TCD having argon as carrier (because I need to detect H2). Since I have both a TCD and an FID installed in the GC, I'm looking now into installing a methanizer and using the FID for CO detection, while leaving the TCD for H2 detection. Moreover, this seems to be the standard configuration in most papers I've read.
I was reading Agilent's manual for the methanizer and it lists the packed column FID (the "universal" version) as a requirement. Unfortunately, I have the "capillary-optimized" version, meaning then that I cannot install a methanizer (without some modifications, at least).
So, the question now becomes: is it possible to convert the capillary column FID to the universal version without having to replace the whole detector? I was looking at the parts list here and it seems to me that only a few parts are different. Therefore, I thought about buying the parts, disassembling the FID, and installing the "packed-version" parts myself.
The new parts would be (according to me):
18710-20119 Jet for packed column (0.018 in. id)
G1531-80580 Base weldments
G1531-00160 FID Seal Insulation
19231-80520 Packed Column 1/8 in. adapter
19234-60715 Nutwarmer insulation
19234-60700 Nutwarmer Cup Assembly
Most parts are relatively inexpensive, except for the base weldment. Still, the price is significantly less than a new detector.
Is this possible, or is it again overly optimistic thinking? Has anyone attempted this?
As always, many thanks in advance for your help.