Advertisement

High CO2, N2 and O2

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Recently, we changed the EI source on our GC/MS system. We did an autotune and everything looked fine. We were able to run some samples on it and get good results. But when we tune the GC/MS today, we found high N2 18.1%, high O2 3.8% and 14% CO2. Is this an air leak? The N2 and O2 ratio is about 5 and I never see CO2 so high before.
Check to see if the column nut is loose. If you install a new ferrule sometimes the graphite/vespel ferrules will shrink when heated and after a few oven cycles you will get a leak where the column enters the analyzer.
The past is there to guide us into the future, not to dwell in.
Hi there it's just a leak most likely at the column transfer line interface tighten it a bit and all should be ok...
3 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry