- Posts: 4
- Joined: Thu May 16, 2013 9:32 pm
Our lab has recently acquired a MSQ mass spec (single quad) from Dionex (now ThermoFisher) as part of a preparative HPLC-CAD-MS system. Due to solubility issues most compounds are dissolved in DMSO and injected straight onto the column without any at-column dilution step. We scan a range from 100-1000 Da in the TIC and are having major problems with DMSO and DMSO adducts in that mass range. Specifically, with the m/z=120, and m/z= 157 peaks corresponding to [DMSO+acn+H]+ and [2*DMSO+H]+. The intensity of these peaks drops of as the DMSO elutes off the column, but the baseline stays significantly elevated due to these peaks even at 15 minutes after the DMSO elution. As a consequence, the analyte peaks appear very weak in the TIC. This is only happening with the MSQ mass spec from ThermoFisher. The mass specs from other manufacturers show the DMSO peaks but at much lower intensities. I think this is effect is due to the geometry of the ion source chamber (since the probe is fixed at a certain angle to the cone) but I am not sure. We use generic gradient runs from 1-99% acetonitrile on C18 columns at flow rates between 35-50 mL/min. Mobile phase = water, acetonitrile (TFA). And no, we cannot get away from DMSO, DMF or NMP as our sample solvents. Any suggestions or insight is appreciated. Thank you.