This is my very first message on this board as my frustration goes bigger. I have tried to find answers on google and on this forum prior to posting, but couldn't find "that" perfect answer. If the answer to this question is very straightforward and i should have found it earlier, then i apologise for it.
I am trying to see my products on an HPLC-RID Agilent Infinity 1260. My column is a NH2 zorbax. I'm using a mixture of CH3CN:H2O, and operate the column at RT, and the detector at 40°C. The flow rate is 1ml/min and injection volume is 10microliters.
The products i have in my mixture are polyols, with a concentration going from 1 to 20 g/L (or mg/mL...). I see very small peaks, that i was able to identify with some standards, but here is the thing:
The peaks are so very small that i am not very confident in the precision(accuracy) of my areas. Plus, some peaks can quite easily be confused with noise. Is there anything i can do, except concentrating my sample, to react a higher intensity of peaks? I'm not sure what the impact of flow rate and/or column temperature would do on this setup. I read that the best temperature for the detector was 5°C above RT, but that is about it.
I am a total novice to HPLC, and even worse for HPLC-RID, and this is why i ask for your help.
I thank you for reading me, have a good day.