Lamp Intensity Test

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hey guys,

Quick question here on calibration of HPLC and failures...

If I am performing a calibration on an Agilent 1100 series HPLC and it fails the low end (190 - 220nm) wavelength check how can I say that my data for the past 6 months (since the last calibration) is still valid when tested at those wavelengths?

Thank you for your assitance,


A
Hey guys,

Quick question here on calibration of HPLC and failures...

If I am performing a calibration on an Agilent 1100 series HPLC and it fails the low end (190 - 220nm) wavelength check how can I say that my data for the past 6 months (since the last calibration) is still valid when tested at those wavelengths?

Thank you for your assitance,


A
Are you looking for valid spectra or just valid results from a chromatographic analysis? If you are looking at a chromatographic analysis and you are doing a calibration check standard each day then that would verify the calculated amount as being valid or failing versus the calibration.

Here we don't use full spectra so the only thing a low lamp check would call into question is sensitivity at those wavelengths, and as long as we can have peaks with s/n >3 we are still passing even if lamp intensity is questionable.
The past is there to guide us into the future, not to dwell in.
Here we don't use full spectra so the only thing a low lamp check would call into question is sensitivity at those wavelengths, and as long as we can have peaks with s/n >3 we are still passing even if lamp intensity is questionable.

This is more or less what I am asking. We have certain methods that run at a wavelengh of 210nm and 220nm. So if my intensity test fails does that mean I need to go back and reevaluate all of the past HPLC runs that used a wavelength in the 190nm - 220nm range.

I'm assuming somewhere in your interal SOPs you have listed the criteria for s/n > 3 = valid even though the intensity test would have failed.
So if my intensity test fails does that mean I need to go back and reevaluate all of the past HPLC runs that used a wavelength in the 190nm - 220nm range.
I would say "no". As long as you passed the system suitability test for a method when you ran it, the system was suitable on that day.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Great point. Thank you for the help.
5 posts Page 1 of 1

Who is online

In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry