internal standard variation
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
I saw 10% CV of internal standards response in seven different calibration standards. Sample preparation is ACN precipitation. According to your experience, is this because of different addition of IS to different calibration standards or just variation of LC-MS in different injections?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
How did the analytes behave? And, more important, how did the ratio if IS to analyte response behave?
If the ratio was more reproducible than the areas, then it's probably an instrument issue or a sample prep issue (after addition of the IS). Correcting for those variations is the reason for using internal standards in first place.
If the ratio was equal or less reproducible than the areas, then variation in the amount of IS is the most likely problem.
If the ratio was more reproducible than the areas, then it's probably an instrument issue or a sample prep issue (after addition of the IS). Correcting for those variations is the reason for using internal standards in first place.
If the ratio was equal or less reproducible than the areas, then variation in the amount of IS is the most likely problem.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
yes, this is very good method to distinguish between the two.
How did the analytes behave? And, more important, how did the ratio if IS to analyte response behave?
If the ratio was more reproducible than the areas, then it's probably an instrument issue or a sample prep issue (after addition of the IS). Correcting for those variations is the reason for using internal standards in first place.
If the ratio was equal or less reproducible than the areas, then variation in the amount of IS is the most likely problem.
3 posts
Page 1 of 1
Who is online
In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 29 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 29 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science