thanx for clearing my doubts in advance
regards
amaryl
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
In principle, yes. In practice, maybe (depending on cis/trans with respect to what: a double bond? a ring?). If reversed-phase does not work, normal-phase often provides better selectivity for isomers.Is it possible to achieve cis trans isomer separation on C18 column.
I'm not sure what you mean by "the physicochemical properties differences of isomers". The method development approach would be the same as for any other separation.Is their a particular approach followed in mobile phase apart from the physicochemical properties differences of isomers...to achieve such separation.
Again, maybe. Depends on where the ionizable group(s) are located with respect to the cis/trans center.their uv spectra is different as stated by sir. do their pka even differ.
If your sample contains only the two isomers, then most likely isocratic will work. Gradient conditions are normally required where you have a wide range of retention (unlikely in the case of only two isomers).is this separation achieved by gradient mode or can even work with isocratic mode.
You don´t have access to a MS?
Does your ghost exceed the original glimepiride (what is it?) in size, same vol and concentration injected?
What happens if you collect the two peaks, separatly, and reinject?