Dexchlorpheniramine maleate [May 11, 2004]
Posted: Wed Aug 25, 2004 3:51 am
by admin
By Renata on Tuesday, May 11, 2004 - 09:15 am:
Hi people!
I'm trying to develop a method for Dexchlorpheniramine maleate. The mobile phase chosen is Phosphate Buffer 20 mM pH 6,2 and Acetonitrile.
Well, the problem is that 2 peaks appear....
one of them is ok with SST, but the other seems to be a great and large mountain.
I thought "maybe, this horrible peak is maleate, and the perfect one, dexchorpheniramine. ". But I made a test with Enalapril maleate (just to check the maleate) using the same conditions and NO GREAT AND LARGE MOUNTAIN, and no PERFECT peak.
I ended up understanding nothing... if it's not maleate maybe is dexchlor, but the first peak isn't maleate either... so dexchlor gives me two peaks????? (many exclamation marks in my mind)
Any light to clarify my ideas??
Thanks (again)
Renata
By Uwe Neue on Thursday, May 13, 2004 - 04:41 pm:
Enalapril can give you broad peaks due to interconversion of conformers during the HPLC run.
There is no reason for this to happen with your chlorpheniramine metabolite. Try to get some pure maleic acid or maleate to see if the other peak is maleic acid.
By Renata on Friday, May 14, 2004 - 09:10 am:
thanks Uwe.
I'll try to get it...
Renata
By Anonymous on Friday, May 14, 2004 - 02:30 pm:
Try pH 3.0, add mobile phase modifier, like triethylamine, you will get maleate peak, early in chromatogram.. active peak should be better now.
By Renata on Monday, May 17, 2004 - 07:47 am:
ok I'll try this one
Thanks a lot.
Renata
By syx_gf on Monday, June 14, 2004 - 12:43 am:
Dear Renata,
Do you know pKa value of dexchlorpheniramine maleate?
Syx
By nisha jacob on Friday, June 18, 2004 - 12:27 am:
whether the enalapril maleate tablets test can be done without placebo, the maleic peak how can confirm
By syx_gf on Monday, June 21, 2004 - 06:13 pm:
We have more than one of API in our formula. I use pH 2.5 and 0.2% TEA on CN-column. The tailing is still between 1.5 - 2.0. We cannot add more TEA to reduce Tf because the peaks of our API will very close.
Any suggestion?
By Anonymous on Monday, June 21, 2004 - 08:49 pm:
Syx,
What do you have in your formula. Can you specify?
By syx_gf on Tuesday, June 22, 2004 - 05:39 pm:
Betamethasone valerate - dexchlorpheniramine maleate
