LC-MS/MS method for polar compounds
Posted: Fri Nov 15, 2013 3:40 pm
by pramodkollu
Hi All,
I got 3 analytes (no structures provided) with little information. Compounds similar to Cytidine monophosphate and one unphosphorylated. Trying to get a LC-MS method for quantitation in biological matrices. Due to polar nature of the compounds, there is no retention of phosphorylated analytes on a regular C18 columns. Do I need to go with ion-pairing chromatography, which I'm trying to avoid if at all possible. I believe the monophosphate is polar acidic compound. Is HILIC an option here? Any suggestions please - Thanks!
Re: LC-MS/MS method for polar compounds
Posted: Sun Nov 17, 2013 7:23 am
by yellaiah2013
bisphosphonate,monophosphonate are not ionise in MS/MS due to their High polarity.eg.Risedronate.
So we should go for derivatives method.
After derivative their molecular weights and properties are chanhed and ionise in MS/MS.
Posted: Tue Nov 19, 2013 5:20 pm
by JMB
This is not correct; phosphonates are detectable by ESI-MS in the negative ionization mode.
Posted: Tue Nov 19, 2013 7:20 pm
by jenche
Hi All,
HILIC might help you. There are some papers/application note that might help you :
- http://www.ncbi.nlm.nih.gov/pubmed/22816250
- http://www.merckmillipore.com/showbrochure/201008.054.ProNet.pdf
Posted: Tue Dec 17, 2013 9:01 pm
by pigfarmer
From my experience you can separate phosphonucleotides on a C18 column (AMP/ADP/ATP) with just isocratic flow rate with say, phosphate buffer.
Posted: Tue Apr 08, 2014 10:34 pm
by ddjchemist
It might not be possible to accurately analyze those compounds in their pure form (without derivatization). The reason is that they are highly polar and they get accumulated on stainless steel (you can find many articles about this topic). Almost everyone neglects this fact.
Good Luck,
Dan