Calculating concentrations help
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
When I inject 5 micro litres and 20 micro litres of the same sample and then calculate the results using my calibration curve, my results tell me that I have more concentration with 20 micro litres. This can't be so as it's the same sample and therefore should be the same concentration - just the amount is different. Can someone please explain.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Difficult to explain without knowing *exactly* what you did.
How did you generate the calibration curve?
Were the units on the amount axis mass or concentration?
Did you use a constant volume with individual concentrations or one concentration and different volumes?
Is the problem sample near the top or bottom of the linear range (or is it in the middle)?
How repeatable is the concentration difference you are seeing?
How much more? 4x ? a few percent?I have more concentration with 20 micro litres
How did you generate the calibration curve?
Were the units on the amount axis mass or concentration?
Did you use a constant volume with individual concentrations or one concentration and different volumes?
Is the problem sample near the top or bottom of the linear range (or is it in the middle)?
How repeatable is the concentration difference you are seeing?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 27
- Joined: Fri May 10, 2013 12:17 am
Concentration does not change of course. The amount on column changes. Example: If your solution is 100 ug/ml, and you inject 0.005 ml (5 micro liters) then 0.5 ug (micrograms) is being injected. If you inject 0.020 ml then 2 ug is being injected. With different volumes you are looking at mg on column only.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
It would've been scary if it wasn't the case i.e. more stuff on column when injecting larger volume 
The solution is simple --> Normalize the result with the injection volume.
Best Regards
The solution is simple --> Normalize the result with the injection volume.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 39
- Joined: Thu Feb 14, 2013 8:54 pm
It could be that you are overloading your column, depending on your concentration. Are you seeing any tailing on the 20 mcL injection? THat can make integration difficult and lead to higher values. It would help for us to see the two chromatograms.
5 posts
Page 1 of 1
Who is online
In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 22 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 22 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science