Shimadzu preparative HPLC - PDA Standby problem

Posted: Wed Nov 13, 2013 9:47 am
by magic_acetone
Hi all,

We have a HPLC preparative shimadzu. We have the LC connected to a PDA module. Everything looks fine, so LC is ready and PDA is also ready. Once we inject, we see that the PDA is changing to Standby modus and from there it does not change once the LC is running. So, that means that we can not get PDA information.

Does anyone knows this problem? I guess that is a communication problem...but is a bit bizard!

Any help would be appreciated!

Best,
Jordi

Re: Shimadzu preparative HPLC - PDA Standby problem

Posted: Wed Oct 08, 2025 9:41 am
by ctlyohanmoreno
HI!, i know it is a long time ago since you posted this, but we are having exactly the same problem, did you solve it, if so, how did you solve it?

Re: Shimadzu preparative HPLC - PDA Standby problem

Posted: Tue Oct 14, 2025 9:12 am
by H.Thomas
Have you checked your method? In the tab "Data Acquisition" the detector should be checked and reasonable acquisition start and end times set.