I'm having some issues with my GC. Here is my setup
Purged Pack inlet (no auto-sampler)
Capillary column (60m or 30m x 0.53 x 0.50)
TCD detector
Chemstation B.04.02 with A.01.14 firmware on Windows XP.
The problem is the same in the front and the back inlet: when I put the system on Pressure Constant Mode and I ask 2mL/min the system gives me 3.074 psi which is correct but it says my Flow is 0mL/min. I take my flowmeter and measure 2mL/min. So in theory I could still use it and get the correct data. The panel is just reading wrong.
When I put the system in Constant Flow Mode and I ask 2mL/min it says it's giving me the right flow but the pressure shoots up to 8psi and the flow that I actually read with my flowmeter is 6mL/min.
I would really rather use the system in Constant Flow but I can't… I've tried dealing with it on my own but I wasn't getting anywhere so I called Agilent and they sent the technician (good thing the GC is still under contract).
We checked for leaks, no leaks
We changes the back EPC, no change
We changes the Logic Board, no change
We upgraded the Firmware to A.01.15 , no change
The next step in changing the power and analog board!
I'm thinking of upgrading my Chemstation to B.04.03 but I'm not convinced that it would help since I have another identical GC on the same computer and its working fine.
Any ideas ?