jenche » Thu Nov 14, 2013 8:20 pm
Nowadays, controlled room temperature is not really important for the UPLC part. Most, if not every, UHPLC systems have a column oven (keeping the column temperature stable, and so the retention times) and a thermostated autosampler (keeping the samples temperature stable).
MS systems require a stable temperature to prevent any calibration shift, that could lead to mass precision deviation during the analyses.
If you use a high resolution mass spectrometer, you really should have a controlled room temperature within a +/-2C range (and prevent any rapid temperature fluctuation, even inside this range).
Also, even if your room temperature is controlled, never install a mass spectrometer near a heat source (or in front of a window without blinds, exposed to direct sunlight), or near the air conditioning system.
And in any case, always prevent temperature exceeding 30°C in the lab to prevent any overheating of the mass spectrometer.