how much pressure normal Septa can handle
Posted: Fri Nov 01, 2013 4:17 pm
by jiangds06
right now, I have around 35psi. Is this gonna cause a problem?
Re: how much pressure normal Septa can handle
Posted: Sat Nov 02, 2013 12:02 am
by dblux_
Nope.
Posted: Mon Nov 04, 2013 2:00 pm
by rb6banjo
35 psig? What column do you have in your GC? For a 30 m x 0.25 mm x 0.25 µm phase that's about 70 cm/s at 40 °C with a helium carrier gas and an atmospheric pressure detector. That's awfully fast.
Posted: Mon Nov 04, 2013 7:38 pm
by Consumer Products Guy
35psi is not that much for a packed column...
Posted: Tue Nov 05, 2013 9:13 am
by thohry
That's no problem. My colleague runs detailed hydrocarbons analyses, and he needs about 100 psi. The column is a 100m capillary.
Posted: Tue Nov 05, 2013 2:21 pm
by rb6banjo
Does he use a hydrogen carrier?
I realize that it may work for him but 80 cm/s (100 psig) is pretty far from the "optimum" flow for a helium carrier gas on a 100 m column. For helium, the van Deemter curve shows the smallest plate heights at about 20-30 cm/s.
Posted: Tue Nov 05, 2013 5:19 pm
by dblux_
In turn I realize that you have not taken into account that these 100 m columns have usually 0,25 mm ID and that type of application is usually not isothermal. What if the final temperature is 250 degC :-) (constant flow).
When assessing septum pressure tightness, maximum pressure during run is the right value.
MfG
Posted: Tue Nov 05, 2013 6:04 pm
by rb6banjo
Ok, if you're starting at 250 °C with that dimension, then the linear velocity is more like 38 cm/s. That's a lot closer to the 20-30 cm/s that is considered optimal for helium.
I'm looking for much smaller molecules than that, hence my statements came from my paradigm.
Posted: Wed Nov 06, 2013 6:46 pm
by James_Ball
35psi should be no problem. If you have a leak detector just check occasionally to make sure you are not leaking through the septa since the needle does slowly wear away the opening, if so then replace and you should be ok.
Posted: Thu Nov 07, 2013 9:43 pm
by jiangds06
Thanks a lot!