Shimadzu LC-10AD Help
Posted: Fri Nov 01, 2013 11:44 am
by Hippi
The pump works, but solvent is not went. What can it be?
Re: Shimadzu LC-10AD Help
Posted: Sun Nov 03, 2013 8:03 pm
by uzman
Possible causes :
Broken piston / pistons ( since LC-10AD has double pistons )
Checkvalves ( especially inlet checkvalves )
Clogged mobile phase filter
The most possible cause is the inlet checkvalves.
Loosen and remove the nut at the outlet checkvalves and try to suck the mobile phase from here with a plastic syringe , then fix the nut and try to pump the solvent.
Re: Shimadzu LC-10AD Help
Posted: Sun Nov 10, 2013 8:07 pm
by artirm
It indeed looks like a check valve problem. Happened to our shimadzu several times, especially when people used phosphate buffers.
Re: Shimadzu LC-10AD Help
Posted: Thu Nov 14, 2013 12:50 pm
by Hippi
Thanks for the advice. I cleaned IN check valve in an ultrasonic bath with isopropyl alcohol and it solved the problem.