I am running some samples at a Shimadzu 2014 GC-FID (100m column) and after some samples some peaks which should be quite small appear significantly bigger, along with some ghost peaks, mainly in the end of the gradient.
Especially the last ghost peak appear with long tailing.
The last stage of the gradient I use is 30 min at 240oC, so I am wondering if high temperature for that long coulld affect the column.
The column, ferrules, liner +o-ring are new and I don;t think that the injector septa is bleeding. Also, the syringe has 3 rinses before every injection and all peaks are eluted since I have no peaks for at least 6 min in the end. The gases I use are also not conatminated.
I also thought that in the end of my sample preparation I have to evaporate the solvent A (isooctane) and redilute it in hexane as it is more friendly to the column, may be a mixed solvent could cause something like this? Though I am positive of the complete evaporation of isoctane.
The same problem appeared in the previous column which was ''older'', so when I changed it to a new one I realized that that peaks were ghost peaks.But this column have only been used for around 100 samples so far.
My concern is not to deteriorate faster the column, since I can avoid integrating the ghost peaks as long as they do not interfere with the peaks I want.
Any suggestions???