NH2 columns for Normal Phase

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

8 posts Page 1 of 1
Did anyone work with NH2 columns on normal phase? What is a typical gradient for NH2 column on NP?

I have problems with a product that break on RP. On NP using silica, the product was retained on the column.

Thanks in advanced
Than I would highly recommend a Standard HILIC column, that will solve your Problem.
Gerhard Kratz, Kratz_Gerhard@web.de
Yes, an amino column can be used in "traditional" normal-phase mode. A typical starting point would be similar to that for silica: a "scouting gradient" starting with 1% isopropanol in hexane and running up to 100% isopropanol. That's more or less the equivalent to the standard 5% - 100% ACN scouting gradient in reversed-phase. Note that the relationship between mobile phase strength and composition in normal-phase is non-linear, so that a disproportionate part of the strength increase occurs in the first 10% of the gradient (but that's true for silica as well).

That said, if your compounds are very polar, and especially if they have ionizable functional groups, then Gerhard's suggestion about HILIC is a good one.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Gerhard, Tom

thank you very much with your solutions.

My first option for NH2 column was HILIC, but it is a non polar compound. That the reason to suggested NP.

Tom, I would try to use Heptane - EthylAcetate in a gradient, what do you think about it? I have never used Isopropanol in our lab on NP.

Thanks in advance
Tom, I would try to use Heptane - EthylAcetate in a gradient, what do you think about it? I have never used Isopropanol in our lab on NP.
No problem. The reason for suggesting IPA is that it is the most polar solvent that is still miscible with hexane (not sure about heptane) and it has a low UV cutoff. EtAc is fine, it just gives you a bit narrower polarity range and a higher UV cutoff.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks Tom

I'll tell you what happen
i'm puzzled: if you're not using reverse phase because your compound isn't stable in reverse phase conditions, why would it survive hilic any better anyway? If you're working on an analyte that degrades rapidly in water/acetonitrile, then it's likely to be a struggle to handle it...
Here is our alternative approach for analysis of polar molecules in hydrogen-bonding mode - SHARC 1 column (Specific Hydrogen-bond Adsorption Resolution Chromatography)

http://www.sielc.com/upload/file/pdf/SHARC_1.pdf
For molecules which are polar and which are not stable in aqueous mobile phase
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
8 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry