I use a RP-C18 column and try to separate only two compounds and need to quantitative these. For the separation I would need an isocratic run (80 % ACN and 20% water).
I have following problems:
1) if I only use 100% ACN the runs are reproducible, the pressure and the retention time are over 10 runs constant but as soon as I add only 10% of water the retention times shift per repeated run although I use 10 column volumes to recalibrate the column. Why does the retention time shift between repeatable injections.
2) When I change from 20% Water / 80 % ACN back to 100 % ACN I see the same effect means I let run the system at least 10-15 Column volumes of with only ACN before I start with the injection of my sample. I do 10 times the same run and after 5 runs I see that the retention time start to be more or less the same although the never go back to the value that I had in the beginning of the day with 100% ACN.
3) Sometimes during the run the pressure drops to 0 bar although it seems that the pump is pumping the solvents in the correct amount. We though of a stuck check valve but they just got replaced. And even if we try to purge with Isoprop the pressure does not come back again. Sometimes it just works for several days and before the same mistake with the pressure starts all over again. What can it be ?
What I am doing wrong?
We thought already that maybe the proportionation of the solvents is inconsistent and the Propvalve does not close or open reliably but we made for all four solvent channels the pumproportionation test and they look ok.
We had just the service guy here to change the check valves and we also replaced all filters and solvents. Why do I see this effect.
If anybody has similar problems or can explain it to me please do so, because I do not have further solutions for it.
I am really glad for any kind of advice or help to solve this problems.