Fluoride Failing low
Posted: Fri Oct 25, 2013 7:20 pm
Hi everyone,
We are having a problem with our Dionex ICS-2000. It use to have a CR- TC Tower, but our calibrations were not passing for any of the anions, so changed it over to a Carb/Bicarb eluent late last year. Essentially it is has become an ICS-1000. Recently, we cannot get the analyte fluoride to pass when we are calibrating the instrument. The ICV (initial calibration validation) for fluoride keeps failing low, all the other anions are passing.
We have AS22/AG22 2 mm columns installed with a 300 ASRS 2mm suppressor. Flow rate is 0.30 mL/min. Pressure and conductivity are at ~1614 psi and ~20.2 µs. We buy the 100x of our eluent pre-made from Dionex/Thermo Scientific.
Here is what we have tried and it has not worked:
1. Changed our calibration standard three times, we can get fluoride to pass on our other instrument, a DX-120, so we know it is not our standard.
2. Added eluent to the vials to get rid of the water dip.
3. Lowered our mA on our suppressor to 50.
4. Both our columns were changed a month ago.
5. Changed our suppressor.
6. Changed the injection valve components.
Would anyone have any other suggestions that we have not tried? We would really like to get fluoride to pass on this instrument again since the run time is 15 minutes compared to 25 minutes on our DX-120.
Thank you
We are having a problem with our Dionex ICS-2000. It use to have a CR- TC Tower, but our calibrations were not passing for any of the anions, so changed it over to a Carb/Bicarb eluent late last year. Essentially it is has become an ICS-1000. Recently, we cannot get the analyte fluoride to pass when we are calibrating the instrument. The ICV (initial calibration validation) for fluoride keeps failing low, all the other anions are passing.
We have AS22/AG22 2 mm columns installed with a 300 ASRS 2mm suppressor. Flow rate is 0.30 mL/min. Pressure and conductivity are at ~1614 psi and ~20.2 µs. We buy the 100x of our eluent pre-made from Dionex/Thermo Scientific.
Here is what we have tried and it has not worked:
1. Changed our calibration standard three times, we can get fluoride to pass on our other instrument, a DX-120, so we know it is not our standard.
2. Added eluent to the vials to get rid of the water dip.
3. Lowered our mA on our suppressor to 50.
4. Both our columns were changed a month ago.
5. Changed our suppressor.
6. Changed the injection valve components.
Would anyone have any other suggestions that we have not tried? We would really like to get fluoride to pass on this instrument again since the run time is 15 minutes compared to 25 minutes on our DX-120.
Thank you