please see image above. Is this too much noise for the GCMS baseline? It seems a lot of column bleeding.
How much noise (in terms of intensity) can be tolerated in the GCMS blank run?
For a perspective on the noise in your system, that's not a very good view. It needs to be more expanded at the baseline.

You need to evaluate this for yourself. Can you detect an analyte of interest at the detection limit. If you can (signal should be 3 times the noise to say it's detectable), then your system is working well for your purpose. If not, then you might have some work to do.
Maybe it's just me, but that picture is too small to determine what, if anything, is happening at the baseline. Can you post something a lot bigger?
Time flies like an arrow. Fruit flies like a banana.
4 posts Page 1 of 1

Who is online

In total there are 38 users online :: 2 registered, 0 hidden and 36 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot], Semrush [Bot] and 36 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry