The negative peaks wheninjecting ultrapure water indicate ions in the eluent.
You will find the retention times identical to e.g. chloride, nitrate, and sulfate.
This means that you add these ions with the eluent components (including the water used for eluent preparation).
It is clear that you see a flat baseline when injecting eluent as you inject the same blank concntrations.
Be aware that your results are influenced by this. If you find e.g. a negative chloride peak corresponding to 1 ppb and you inject a sample of 2 ppb the peak you get will correspond to 1 ppb only!
I see. Thanks for the clarification.
And yes, those retention time are identical with my standards.
I've got a few questions regarding it:
1. If that's the case as you mentioned, does that indicated that there are ions
contamination in my ultrapure water system? Or maybe somewhere in the IC system?
2. I'm thinking of using eluent as my blank, and also using the eluent to do dilution on my standards. Is it be okay?
If the eluent itself showed flat baseline, probably it won't influence much on my results, right?
(Because if my ultrapure water is the main cause of contamination, it'll interrupting my result by showing negative peak in the blank; but not for the case of eluent).
Thanks again.