krickos » Thu Oct 24, 2013 11:08 am
Hi Below is some text cut out from <621> USP.
I get the feeling that your are reducing particle size too much ie more than 50%, which would triggger a formal validation that would likely need approval by FDA (alternativly filing of an alternative procedure to USP).
Reducing paticle size stepwise may seem neat but you would still be outside the 50% limit as the dimension in the monograph decides, not what you reduced to in a prior method.
As I see it you can go from 10 to 5µm, 5 to 3,5µm or 3,5 to 2,1µm without tempering with the 50% limit.
To my knowledge USP is thinking on a chapter on how to convert HPLC procedure to UPLC, to facilitate a shift to smaller columns/particles without excessive revalidation and regulatory approval.
My point being if you operate within what USP allows and you refer to USP in your file, you do not have to notify FDA or similar if you operate outside there is a risk you have to be aware of or additional work needed to get it approved.
Regards Chris
Adjustments to chromatographic systems performed in order to comply with system suitability requirements are not to be made in order to compensate for column failure or system malfunction.
If adjustments of operating conditions are necessary in order to meet system suitability requirements, each of the items in the following list is the maximum variation that can be considered, unless otherwise directed in the monograph; these changes may require additional validation data. To verify the suitability of the method under the new conditions, assess the relevant analytical performance characteristics potentially affected by the change. Multiple adjustments can have a cumulative effect on the performance of the system and are to be considered carefully before implementation. Adjustments to the composition of the mobile phase in gradient elution are not recommended. If adjustments are necessary, only column changes (same packing material) or dwell volume adjustments are recommended.
pH of Mobile Phase (HPLC): The pH of the aqueous buffer used in the preparation of the mobile phase can be adjusted to within ±0.2 units of the value or range specified.
Concentration of Salts in Buffer (HPLC): The concentration of the salts used in the preparation of the aqueous buffer employed in the mobile phase can be adjusted to within ±10% if the permitted pH variation (see above) is met.
Ratio of Components in Mobile Phase (HPLC): The following adjustment limits apply to minor components of the mobile phase (specified at 50% or less). The amounts of these components can be adjusted by ±30% relative. However, the change in any component cannot exceed ±10% absolute (i.e., in relation to the total mobile phase). Adjustment can be made to one minor component in a ternary mixture. Examples of adjustments for binary and ternary mixtures are given below.
Binary Mixtures
specified ratio of 50:50: 30% of 50 is 15% absolute, but this exceeds the maximum permitted change of ±10% absolute in either component. Therefore, the mobile phase ratio may be adjusted only within the range of 40:60 to 60:40.
specified ratio of 2:98: 30% of 2 is 0.6% absolute. Therefore the maximum allowed adjustment is within the range of 1.4:98.6 to 2.6:97.4.
Ternary Mixtures
specified ratio of 60:35:5: For the second component, 30% of 35 is 10.5% absolute, which exceeds the maximum permitted change of ±10% absolute in any component. Therefore the second component may be adjusted only within the range of 25% to 45% absolute. For the third component, 30% of 5 is 1.5% absolute. In all cases, a sufficient quantity of the first component is used to give a total of 100%. Therefore, mixture ranges of 50:45:5 to 70:25:5 or 58.5:35:6.5 to 61.5:35:3.5 would meet the requirement.
Wavelength of UV-Visible Detector (HPLC): Deviations from the wavelengths specified in the procedure are not permitted. The procedure specified by the detector manufacturer, or another validated procedure, is used to verify that error in the detector wavelength is, at most, ±3 nm.
Stationary Phase
column length (gc, hplc): Can be adjusted by as much as ±70%.
column Inner diameter (hplc): Can be adjusted if the linear velocity is kept constant. See Flow Rate (HPLC) below.
Particle Size (HPLC): The particle size can be reduced by as much as 50%, but cannot be increased.
Flow Rate (HPLC): When column dimensions have been modified, the flow rate can be adjusted using:
"picture of formula missing"
in which F1 is the flow rate indicated in the monograph, in mL/min; F2 is the adjusted flow rate, in mL/min; l1 is the length of the column indicated in the monograph; l2 is the length of the column used; d1 is the column inner diameter indicated in the monograph; and d2 is the internal diameter of the column used. Additionally, the flow rate can be adjusted by ±50%.
Injection Volume (HPLC): The injection volume can be reduced as far as is consistent with accepted precision and detection limits; no increase is permitted.
Column Temperature (HPLC): The column temperature can be adjusted by as much as ±10. Column thermostating is recommended to improve control and reproducibility of retention time.