lmh » Wed Oct 23, 2013 9:51 am
hmmm, it's always difficult to deal with a literature method that isn't quite working.
(1) The low percentage of acetonitrile in B. Actually this doesn't worry me. If a method only needs 10% ACN, what's the problem? If you want to pump a gradient from 0-10% accurately, you will get a better gradient by making 10% ACN by hand and pumping 0-100% than by using 100% ACN and pumping 0-10%. The only downsides are that you don't have a big reserve of organic available for washing undesirable strongly retained things from the column, and you need to mix solutions yourself.
(2) But not all C18 columns like running in 0% organic ("phase collapse"). I have no idea about the column this paper uses.
(3) pH, yes, 4.0 isn't a very sensible place for a phosphate buffer because the relevant pKa values are about 2, 6.8 (and 12.5), so 4 is too far from both the 2 and 6.8 pKa values of phosphate for the result to be a buffer. But if you are considering changing it, you have to think about what will happen to your analytes too. ATP has pKa values at 4, 4.1 and 6.5, so changing from 4 to 7 for the sake of the phosphate will affect the ATP fairly drastically. It will be much more ionised, which will increase its interactions with an ion pair reagent, and decrease what little chance it had of interacting directly with C18. On the other hand, running a column with the pH very close to the pKa of the analyte is usually not a good idea, because it means that minor changes in pH when preparing the buffer can lead to big changes in the chromatography. Also minor variations in pH caused by temperature differences between the middle of a column and the outside of the column can lead to different migration rates, leading to very broad peaks (I've heard... obviously I haven't been measuring the pH at different points in my columns!).
(4) Why on earth the change of pH and change of concentration of buffer? I'd guess that the original author may have been reasoning that an ion-pair method is actually a hybrid of reverse phase and ion-exchange, because the analyte is attached to the pair reagent by an ion-exchange mechanism, and the pair reagent is attached to the column by a reverse phase mechanism. Therefore, the thinking might be, it doesn't matter whether I elute my analyte-ion-pair combined thing off the column by increasing organic, or elute my analyte off the ion-pair reagent by increasing saltiness (or changing pH). I'm not 100% convinced, particularly on the pH, which sort of goes the wrong way. Maybe the saltiness was supposed to elute, and the pH was a mere accident of adding the same tetrabutyl ammonium to two different concentrations of the same phosphate "buffer"? I'd need to make it up to see.