HP 5972 MSD - Fault 512 - Detector HV Supply
Posted: Fri Oct 18, 2013 10:10 pm
I am experiencing a fault 512 with my HP 5972 MSD. The fault results in the message “Difficulty in the Detector HV Supply”. I am wanting to confirm what replacement part I should consider first? The DC power supply, the power distribution board, or something else? If anyone has any thoughts, please let me know. I have checked all the connections leading to the electron multiplier and everything seems fine. All connections are also secure to the analyzer and the vacuum level is below 100 mTorr.
I’ve attempted to do the air/water check several times now in the diagnostics/vacuum section of ChemStation. Every time, I have been unable to turn on the detector due to the HV supply error, except for one time that resulted in the error “Difficulty in the mass filter electronics”. I’d like to start troubleshooting first with the detector HV supply error and move forward from there.
Thanks for any suggestions, I appreciate it!
I’ve attempted to do the air/water check several times now in the diagnostics/vacuum section of ChemStation. Every time, I have been unable to turn on the detector due to the HV supply error, except for one time that resulted in the error “Difficulty in the mass filter electronics”. I’d like to start troubleshooting first with the detector HV supply error and move forward from there.