That is exactly what I see after I run a set! I was also very happy that it washed off quite easily, I have to admit - I was quite scared when I first saw how dirty it wasspray chamber smelt like a toffee-apple and was covered in brown caramel
Yes we have a constant Cone Gas flow on our instrument, we are actually considering increasing the flow as we are having issues with contamination inside mass analyzer (quads & coll cell). Current flow is set at 150L/hr, how does that compare to your setting?As long as you have a good purge flow between the cones(I assume the other instruments have this also) it should keep out most of the non-ionized material which will just build up on the outer cone surfaces.
Yes we have a constant Cone Gas flow on our instrument, we are actually considering increasing the flow as we are having issues with contamination inside mass analyzer (quads & coll cell). Current flow is set at 150L/hr, how does that compare to your setting?[/quote]As long as you have a good purge flow between the cones(I assume the other instruments have this also) it should keep out most of the non-ionized material which will just build up on the outer cone surfaces.
The Waters tech told me the same thing, use highest flow rate without loosing too much sensitivity. Interesting that AB Sciex doesn't have units on the flow, would be nice to compare flows between instruments.I am not sure what the flow is from our API3200 it only list the flow a 5, 10, 20, ect with no units. The tech from AB Sciex told me to set it to the highest setting I could without loosing too much sensitivity and that would keep the insides cleaner.
Yes PSA would help retain sugars, but since this is only a short study and I don't expect it to become a regular one we are not looking into it. But if this does become a regular study, I would defintely try PSA and see how it affects the recovery of our target analyte.Hello; I haven't read all the comments posted here; but I remember that Quechers use PSA in order to retain Sugars; so that could be a way to remove sugar using PSA or Amino sorbentconbined with C18; I do not know if this could affect the recovery of your analyte.
Waters instrument has this as well, need to ensure its set up for this new method - I completely forgot about it, thank you!I don't know the Waters Instrument, but our Agilent has a switch valve. I usually pump the first 1-2min of each run into waste before switching to the ESI when analyzing sugar-containing samples.