Agilent 1290(new), M.Phase: 0.5%formic acid in H2O,PH8.5 by NH3/MeCN
DAD:230,4 ref:300,100
By the way, why when we use needle wash the vial must be uncapped?
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Sorry to be a stickler, but *which* Prontosil C18, specifically? Which Phenomenex column? Do you know the pH limits of each?i use prontosil c18 250x4.6 ,5μm but i tried one from phenomenex as well with the same results.I "play" with the temperature a bit but it's from 20 to 35C.
it must be the mobile phase. When i wash H2O/ACN: 50/50 the noise is fine. when i put M.phase A in water's channel and begin to equilibrate the noise is about 1mAU (only at 230nm). Formic acid maybe?
i changed channels several times. i don't believe PH 8.5 (+ %ACN) will be much trouble for these columns. the 1 of 2 prontosil i used was brand new.it is long -term noise and it's only for 230nm. In this analysis i have dual wavelength 230/320 and in 320 the noise is fine. This is a brand new LC, i wouldn't blame the lamp. The pressure is pretty steady.I don't blame the glassware either. i have done the same analysis 3 times in 30 days and i had the same problem.