Waters 486 tunable UV detector

Posted: Wed Oct 16, 2013 4:22 pm
by yshurubo
Hello, everyone. I have a problem with Waters 486 tunable UV detector. This detector was fine yesterday, but this morning it shows '7FF2' error code.
All knobs are disabled. This error code is not listed in the manual... Any suggestions? Thanks

Re: Waters 486 tunable UV detector

Posted: Tue Oct 22, 2013 12:47 pm
by danko
I guess you've powered the detector down between "yesterday" and "today".
If so, I'd suspect the backup battery. Have you ever replaced it? You might need to replace it now.