I have read in books, that some times gradient runs are insrtument specific. Is it true? Why?

Regards
Ashvin
Ashvin Panchal
The reason has to do with the physical and mechanical configuration of each system. It depends on where in the system the solvents are mixed to form the gradient. Some systems use a separate pump for each mobile phase and the mixing takes place in a mixing chamber or at a "tee". Other systems mix the mobile phases at low pressure using a proportioning valve and them pump the mixed phase at high pressure through the column using a single pump. In every system, there is a certain amount of tubing between the point where the solvent is mixed and the column and this tubing volume (also including the injector) is different for every system. The result is that there is a time/volume delay from the point where the gradient is formed and actually reaches the column.
2 posts Page 1 of 1

Who is online

In total there are 202 users online :: 0 registered, 0 hidden and 202 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 202 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry