ChemStation
Posted: Mon Oct 07, 2013 5:49 pm
by brendandamm
This is probably one of the more dumber questions that is posted on this forum but I am STRUGGLING with the chemstation software. How can I get it to generate a report with every peak listed and their respective k', selectivity, efficiency and resolution? I have the performance report style selected and I have the column installed on the main screen but everytime I generate a report it only gives me the values for the compounds I have calibrated. So I can only really view the data for two peaks at the moment and some of data is incomplete. I hate ChemStation and this is killing me. Please help. I need the report to list every peak in the chromatogram, not just the calibrated compounds. Thanks
B
Re: ChemStation
Posted: Tue Oct 08, 2013 5:57 am
by dblux_
In report options find something like Report Layout for Uncalibrated Peaks.
Re: ChemStation
Posted: Tue Oct 08, 2013 1:54 pm
by Mike H.
We used to use Chemstation. To get these parameters I believe you must first name the peaks in the calibration table and then select the right report template. I think the one you want is extended performance.
Re: ChemStation
Posted: Tue Oct 08, 2013 3:11 pm
by Consumer Products Guy
I've gone from being somewhat expert on Chemstation A revisions to a rookie with Chemstation C revision....maybe somewhat because we never had revision B to stumble with....
Re: ChemStation
Posted: Wed Oct 09, 2013 1:18 pm
by brendandamm
Thanks guys. I feel like I have tried everything. The two main peaks of interest I have calibrated in my calibration table and I get results for those. But I want to be able to view the data for every peak that shows up in the chromatogram and that is what I'm struggling with. ChemStation SUCKS
Re: ChemStation
Posted: Thu Oct 10, 2013 6:46 pm
by Mike H.
You have to name the peaks to get the results you want. Every single one.
Re: ChemStation
Posted: Thu Oct 17, 2013 10:02 pm
by James_Ball
Maybe just make a bunch of analytes called "Unknown1, Unknown2, ect" in the calibration table junk calibration data in them. Then when you have peaks you want the k ' values for you just set their retention times to match up with an unknown and it should then print the extended report values on the report(minus a true concentration that is).