- Posts: 2
- Joined: Tue Oct 22, 2013 9:18 am
What could be happening? Is the compound accumulating in the column?
I tried revese flusing to the column with methanol, IPA, ACN, acetone at a flow rate of 1ml/min, but to no avail.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Is there any marked back pressure difference now and before when you used to get all the peaks? If the pressure is on the higher side then the column maybe clogged. You can try regenerating the column by flushing the column from Polar to non polar solvents like flushing with water followed by Methanol then Isopropanol and finally with Dichloromethane and then follow the sequence in reverse order till water.I have already tried these trouble shooting techniques.
Can the compound be accumulating? Because immediately after running the compound I injected 3ppm pyrene and it did not give me a peak like it had when injected prior to the compund, at 333nm. In fact, the peak disappeared and the peak obtained at 254nm for pyrene was not acceptable.
Any other suggestions?
Is it possible to revive a column by removal of clogging factors, if so, how? Can anyone forward any references, where instead of replacing the column, it has been revived?
It can be injector as well.If you didn’t see a peak, then it’s the detector you are having trouble with.
Best Regards