2 posts Page 1 of 1
Dear all,

I would like to ask you some questions.
I am checking GC DANI 1000 and having some problems with this GC DANI 1000. Please kindly advise:

1. When booting the GC in the beginning, the system notifies that there is error in auxiliary gas (nitrogen) of FID. And when I change the flow rate of this auxiliary gas, the system doesn't apply.

2. How long for the ECD lifetime? And how to check whether ECD detector working or not ?

Thanks in advance for your advise !
Too little info. on the GC. Electronic control of carrier/fuel gases. Sounds like an electrical issue.

ECD The half life of the Ni63 is largely irrelevant. Issues are contamination and electrical; almost always contamination. I can't remember if the 1000 gives and read backs of current flow but the best way to test is to get the GC running and run a test standard.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry