Plate number dependence on extracolumn volume
Posted: Thu Sep 26, 2013 3:55 pm
N = 5.545 (Tr/Wh)^2 is a formula given, e.g., by IUPAC, for calculating plate number from a chromatogram resulting from an isocratic (or isothermal) elution. By it, one has but to measure peak width at half height (Wh) and retention time from time zero (Tr).
I am bothered by this formula (and a related one using base peak width). Because the retention time is measured with time zero being the injection time, rather than with time zero being the elution time of a completely nonretained compound. N therefore will not be independent of the extracolumn volume of the system.
One implication of this is that a column manufacturer could artificially inflate their reported number of theoretical plates, for instance in liquid chromatography by performing their chromatography on a system having a longer length of tubing between the injector and the column head. OK, sure, such a system might also cause an increase in peak width Wh, which would have an opposite effect on (would reduce) plate number, but I'd hardly expect these two effects to cancel each other out completely in all cases. For instance, the longer tube could be made a highly curved (sinusoidally or tortuously curled) one, which configurations are reported to induce less dispersion per length of tubing. Or the sample could be dissolved in a much weaker solvent, allowing it to become refocussed (stack) at the column head.
Am I missing something? Why is plate number such a gold standard for column efficiency?
I am bothered by this formula (and a related one using base peak width). Because the retention time is measured with time zero being the injection time, rather than with time zero being the elution time of a completely nonretained compound. N therefore will not be independent of the extracolumn volume of the system.
One implication of this is that a column manufacturer could artificially inflate their reported number of theoretical plates, for instance in liquid chromatography by performing their chromatography on a system having a longer length of tubing between the injector and the column head. OK, sure, such a system might also cause an increase in peak width Wh, which would have an opposite effect on (would reduce) plate number, but I'd hardly expect these two effects to cancel each other out completely in all cases. For instance, the longer tube could be made a highly curved (sinusoidally or tortuously curled) one, which configurations are reported to induce less dispersion per length of tubing. Or the sample could be dissolved in a much weaker solvent, allowing it to become refocussed (stack) at the column head.
Am I missing something? Why is plate number such a gold standard for column efficiency?