no signal with LCQ classic
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
2 posts
Page 1 of 1
We do not detect any peaks with our LCQ Classic. The diagnostics all check out except for the multipole 1 offset. We have tried cleaning the mass analyzer and stack but the reading remains around -132 although it is set at -3. Any ideas on what could be causing the fault?
-
- Posts: 179
- Joined: Thu Jul 07, 2005 12:38 pm
If I remember right all optics are controlled by the big PCB on the right side.. Not much to due expect exchange the PCB if that is the case.. You could be lucky that it is a power supply but than I would expect other optics to act strange also...
Sorry..
Sorry..
Kind regards
Leadazide
Leadazide
2 posts
Page 1 of 1
Who is online
In total there are 88 users online :: 1 registered, 0 hidden and 87 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot] and 87 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot] and 87 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science