Varian 1200L auto-tune, lens not working properly
Posted: Tue Sep 24, 2013 12:55 pm
Hi
I am working on a Varian 1200L MS/MS instrument (coupled with Varian CP-3800 GC). When I start performing an auto-tune (FC-43 positive EI) on this instrument, the initial hardware diagnostics shows a window with a warning 'Lens 3 is not working properly'. When I choose to proceed with auto-tune anyway, the tuning is successfully completed.
However the ratio of m/z 28 : 18 is much too high (percentages of 100% for 28 and 6.1 % for 18, which makes the ratio of 16.4 : 1, and I found that the limit is 1.2 : 1). Therefore, I need to check the instrument for possible leaks. I do not know too much about it but probably it is a significant leak isn't it?
Coming back to the lens. Probably the warning is a separate issue from the leak. What can I do in this matter? Can I somehow find out in exactly WHAT way it is not working properly? What may be the consequences of working on the instrument with one of the lenses not working properly? I will be very grateful for any comments and help.