strange autosampler behaviour
Posted: Mon Sep 23, 2013 7:53 pm
I had issues with inacceptable low precision om one of our RP-Methods running on an Ultimate system. I think i pinned it down to the autosampler.
During sample stability studies, standard solutions and samples were injected every 6 hours from the same vials.
On positions A3 and A4 as well as C4 peak areas went down by about 2%, while on A8 , B7 and D8 peak areas went up byabout 3%. Some of them are double injections with excellent agreement.
The same behaviour was reproducible on another day.
On a second system there was no clear relationship between position and trends, the RSD however were worse.
I reduced draw.speed from 20 to 3 and increased draw.delay from 3000 to 5000 (25 µl injection) and changed the autosampler solvent to mobile phase composition.
On the first HPLC differences decreased, i have however still vial positions that get higher peak areas (b3 and b8 IIRC) on the second system there was not so much change: the deviation were just slightly smaller.
The method is on two APIs and the effect are abserved for both. the tray is cooled down to 5°C.
Any similar obsevations or ideas?
Alex
Alex