We are in the job to develop a HPLC method for one of API by using following conditions.
Column : Kromasil Eternity C18 250 X 4.6mm,5µm.
Flow rate : 1.0 mL/min
Wavelength : 215 nm
Column temperature : 50ºC
Run time : 37 min
Injection volume : 10 L
Diluent : Acetonitrile: Water (50:50) v/v
Elution : Gradient
Sample concentration : 2.0 mg/mL
Buffer preparation : Take accurately about 3.48gram of Di potassium hydrogen phosphate in1000 mL of Milli-Q-water and filter this solution, through 0.22µm Nylon membrane filter paper.
Mobile phase–A : Buffer: Methanol (95:05)v/v
Mobile phase–B : Acetonitrile:Methanol(80:20)v/v
GRADIENT PROGRAM
TIME(In min) Mobile Phase-A (%) Mobile Phase-B (%)
0.01 55 45
20 30 70
30 30 70
30.1 55 45
37 55 45
Problem-1: When we using the new HPLC column for analysis no impurities observed at non polar side around 26 minutes, after around 60 injections non polar impurities observed around 0.12% (These impurities are matches with base degradation impurities and conformed as dimmers of our API), after 100 injections these impurities raised up to 0.5%, by injecting same sample. Why this behavior was observed and how can we prevent this behavior. No problem with solution stability. This is observed only different sample preparations by different time intervals with same homogeneous sample.
Mobile phase pH is critical for this analysis always we are using basic pH buffer only, other wise good peak shapes can not be observed.
Any one please help me.
Regards
sreenivas