Impurity peak in WORKING STANDARD

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I see an impurity peak in the working standard. System suit and all other criteria pass. What is the limit of imurities that I can ignore or what is the maximum area of the impurity that can be disregarded according to USP?

Thanks in advance
What does your SOP say? :twisted:

Ph. Eur. suggests a general disregard limit of 0.05% (chapter 2.2.46) if I recall correctly.
Thanks for the reply. SOP does not comment anything on impurity in WOrking standard. However it does limit amount of impurity in the sample.
3 posts Page 1 of 1